Improved reproducibility of diffusion MRI of the human brain with a four‐way blip‐up and down phase‐encoding acquisition approach

Link to paper

Improved reproducibility of diffusion MRI of the human brain with a four‐way blip‐up and down phase‐encoding acquisition approach

M. Okan Irfanoglu, Neda Sadeghi, Joelle Sarlls, Carlo Pierpaoli

Abstract

Purpose

To assess the effects of blip‐up and ‐down echo planar imaging (EPI) acquisition designs, with different choices of phase‐encoding directions (PEDs) on the reproducibility of diffusion MRI (dMRI)‐derived metrics in the human brain.

Methods

Diffusion MRI data in seven subjects were acquired five times, each with five different protocols. The base design included 64 diffusion directions acquired with anterior‐posterior (AP) PED, the first and second protocols added reverse phase‐encoded urn:x-wiley:07403194:media:mrm28624:mrm28624-math-0001 posterior‐anterior (PA) PED images. The third one included 32 directions all with PED acquisitions with opposite polarity (AP and PA). The fourth protocol, also with 32 unique directions used four PEDs (AP, PA, right‐left (RL), and left‐right (LR)). The scan time was virtually identical for all protocols. The variability of diffusion MRI metrics for each subject and each protocol was computed across the different sessions.

Results

The highest reproducibility for all dMRI metrics was obtained with protocol four (AP/PA‐RL/LR, ie, four‐way PED). Protocols that used only urn:x-wiley:07403194:media:mrm28624:mrm28624-math-0002 for distortion correction, which are the most widely used designs, had the lowest reproducibility.

Conclusions

An acquisition design with four PEDs, including all DWIs in addition to urn:x-wiley:07403194:media:mrm28624:mrm28624-math-0003 images should be used to achieve high reproducibility in diffusion MRI studies.