Magnetization‐prepared GRASP MRI for rapid 3D T1 mapping and fat/water‐separated T1 mapping

link to paper

Magnetization‐prepared GRASP MRI for rapid 3D T1 mapping and fat/water‐separated T1 mapping

Li Feng, Fang Liu, Georgios Soultanidis, Chenyu Liu, Thomas Benkert, Kai Tobias Block, Zahi A. Fayad, Yang Yang

Abstract

Purpose

This study aimed to (i) develop Magnetization‐Prepared Golden‐angle RAdial Sparse Parallel (MP‐GRASP) MRI using a stack‐of‐stars trajectory for rapid free‐breathing T1 mapping and (ii) extend MP‐GRASP to multi‐echo acquisition (MP‐Dixon‐GRASP) for fat/water‐separated (water‐specific) T1 mapping.

Methods

An adiabatic non‐selective 180° inversion‐recovery pulse was added to a gradient‐echo‐based golden‐angle stack‐of‐stars sequence for magnetization‐prepared 3D single‐echo or 3D multi‐echo acquisition. In combination with subspace‐based GRASP‐Pro reconstruction, the sequence allows for standard T1 mapping (MP‐GRASP) or fat/water‐separated T1 mapping (MP‐Dixon‐GRASP), respectively. The accuracy of T1 mapping using MP‐GRASP was evaluated in a phantom and volunteers (brain and liver) against clinically accepted reference methods. The repeatability of T1 estimation was also assessed in the phantom and volunteers. The performance of MP‐Dixon‐GRASP for water‐specific T1 mapping was evaluated in a fat/water phantom and volunteers (brain and liver).

Results

ROI‐based mean T1 values are correlated between the references and MP‐GRASP in the phantom (R 2 = 1.0), brain (R 2 = 0.96), and liver (R 2 = 0.73). MP‐GRASP achieved good repeatability of T1 estimation in the phantom (R 2 = 1.0), brain (R 2 = 0.99), and liver (R 2 = 0.82). Water‐specific T1 is different from in‐phase and out‐of‐phase composite T1 (composite T1 when fat and water signal are mixed in phase or out of phase) both in the phantom and volunteers.

Conclusion

This work demonstrated the initial performance of MP‐GRASP and MP‐Dixon‐GRASP MRI for rapid 3D T1 mapping and 3D fat/water‐separated T1 mapping in the brain (without motion) and in the liver (during free breathing). With fat/water‐separated T1 estimation, MP‐Dixon‐GRASP could be potentially useful for imaging patients with fatty‐liver diseases.