3T 31P/1H calf muscle coil for 1H and 31P MRI/MRS integrated with NIRS data acquisition

link to paper

3T 31P/1H calf muscle coil for 1H and 31P MRI/MRS integrated with NIRS data acquisition

Bei Zhang, Daniel Lowrance, Manoj Kumar Sarma, Miles Bartlett, David Zaha, Michael D. Nelson, Anke Henning

Abstract

Purpose

Our aim was to design and build a 3T 31P/1H calf coil that is capable of providing both good 31P and 1H transmit and receive performance, as well as being capable of accommodating a near-infrared spectroscopy (NIRS) device for simultaneous NIRS data and MRI/MRS acquisition.

Method

In this work, we propose a new 3T 31P/1H birdcage combination design consisting of two co-centrically positioned birdcages on the same surface to maximize transmit efficiency and sensitivity for both nuclei. The 31P birdcage is a high-pass birdcage, whereas the 1H birdcage is a low-pass one to minimize coupling. The diameter of the 31P/1H birdcage combination was designed to be large enough to accommodate a NIRS device for simultaneous NIRS data and MRI/MRS acquisition.

Results

The one-layer coil structure of the birdcage combination significantly streamlines the mechanical design and coil assembly process. Full-wave simulation results show that the 31P and 1H are very well decoupled with each other, and the 1H and 31P SNR surpasses that of their standalone counterparts in the central area. Experiment results show that the inclusion of a NIRS device does not significantly affect the performance of the coil, thus enabling simultaneous NIRS and MRI readouts during exercise.

Conclusion

Our findings demonstrate the feasibility and effectiveness of this dual-tuned coil design for combined NIRS and MRS measurements, offering potential benefits for studying metabolic and functional changes in the skeletal muscle in vivo.