A z-gradient array coil with a dedicated active-shielded array coil for MRI

link to paper

A z-gradient array coil with a dedicated active-shielded array coil for MRI

Manouchehr Takrimi, Ergin Atalar

Abstract

Purpose

An array-based z-gradient coil with a set of programmable power amplifiers can outperform a conventional z-gradient coil and make it highly customizable with a broader range of tunable features.

Methods

A dynamically adjustable imaging volume can be achieved using a pair of independent arrays and a modified optimization procedure based on analytic equations. Two modes of operation are provided: (a) standard mode that resembles a conventional coil; (b) advanced mode, where all performance parameters can be adjusted employing a controllable feeding mechanism. Commercial software is used to demonstrate the validity and feasibility of the proposed coil.

Results

Primary and shield array diameters are 24 and 30 cm, both of which comprise 12 bundles of 10 turns copper wires. Maximum feeding voltage/current is 250 V/100 A for all array elements. Four distinct magnetic profiles are provided: (a) conventional profile with 140 mm diameter spherical region of interest, 120 mT/m gradient, and up to 4500 T/m/s slew rate; (b) profile of 200 mT/m, 70 mm region of interest, and up to 6900 T/m/s slew rate; © 60 mm axially shifted 70 mm region of interest with 120 mT/m strength and 3600 T/m/s slew rate; and (d) profile of 370 mT/m, 120 mm region of interest, and 3700 T/m/s slew rate when the active shield is reverse fed.

Conclusion

By using an active-shielded gradient array coil, the magnetic field profile of the imaging volume can be adjusted dynamically, and it can provide new features and a wide range of field profiles for diverse applications in MRI.