link to paper
Acoustic noise reduction for spiral MRI by gradient derating
Zeyu Zhou, Abdulrahman Alfayad, Tzu Cheng Chao, James G. Pipe
Abstract
Purpose
To show that the acoustic noise of spiral MRI can be reduced by derating the gradients with minimal penalty to image quality and scan time, and to illustrate an algorithm for optimal choice of derating parameters.
Theory and Methods
Acoustic noise level was measured and compared for various values of maximum gradient amplitude and slew rate for T1-weighted spin-echo spiral scans while maintaining image contrast, FOV and resolution, and readout time. A full gradient trajectory and a derated gradient (undersampled) trajectory were chosen for a volunteer scan followed by parallel imaging-aided reconstruction to illustrate comparable image SNR. Two auto-derating methods, which prioritize slew rate and gradient amplitude, respectively, were derived using analytical results from the WHIRLED PEAS variant of spiral waveforms and compared in their acoustic noise level under test use cases.
Results
Derating the gradients made the scan quieter by 16.6 dB(A) on average than a full gradient trajectory and required an undersampling factor R = 2 in order to maintain scan time, with no appreciable penalty in image SNR. Prioritizing reduced slew rate resulted in maximal loudness reduction.
Conclusion
Scanner gradients can often be derated to reduce the acoustic noise for spiral MRI with minimal penalty in scan time and image quality with the help of parallel imaging. An automatic slew-priority derating method that maximizes loudness reduction is given.