Concomitant magnetic-field compensation for 2D spiral-ring turbo spin-echo imaging at 0.55T and 1.5T

link to paper

Concomitant magnetic-field compensation for 2D spiral-ring turbo spin-echo imaging at 0.55T and 1.5T

Zhixing Wang, Rajiv Ramasawmy, Xue Feng, Adrienne E. Campbell-Washburn, John P. Mugler III, Craig H. Meyer

Abstract

Purpose

To develop 2D turbo spin-echo (TSE) imaging using annular spiral rings (abbreviated “SPRING-RIO TSE”) with compensation of concomitant gradient fields and B 0 inhomogeneity at both 0.55T and 1.5T for fast T2-weighted imaging.

Methods

Strategies of gradient waveform modifications were implemented in SPRING-RIO TSE for compensation of self-squared concomitant gradient terms at the TE and across echo spacings, along with reconstruction-based corrections to simultaneously compensate for the residual concomitant gradient and B 0 field induced phase accruals along the readout. The signal pathway disturbance caused by time-varying and spatially dependent concomitant fields was simulated, and echo-to-echo phase variations before and after sequence-based compensation were compared. Images from SPRING-RIO TSE with no compensation, with compensation, and Cartesian TSE were also compared via phantom and in vivo acquisitions.

Results

Simulation showed how concomitant fields affected the signal evolution with no compensation, and both simulation and phantom studies demonstrated the performance of the proposed sequence modifications, as well as the readout off-resonance corrections. Volunteer data showed that after full correction, the SPRING-RIO TSE sequence achieved high image quality with improved SNR efficiency (15%–20% increase), and reduced RF SAR (˜50% reduction), compared to the standard Cartesian TSE, presenting potential benefits, especially in regaining SNR at low-field (0.55T).

Conclusion

Implementation of SPRING-RIO TSE with concomitant field compensation was tested at 0.55T and 1.5T. The compensation principles can be extended to correct for other trajectory types that are time-varying along the echo train and temporally asymmetric in TSE-based imaging.