Efficient optimization of chemical exchange saturation transfer MRI at 7 T using universal pulses and virtual observation points

link to paper

Efficient optimization of chemical exchange saturation transfer MRI at 7 T using universal pulses and virtual observation points

Thaddée Delebarre, Vincent Gras, Franck Mauconduit, Alexandre Vignaud, Nicolas Boulant, Luisa Ciobanu

Abstract

Purpose

To optimize the homogeneity of the presaturation module in a Chemical Exchange Saturation Transfer (CEST) acquisition at 7 T using parallel transmission (pTx).

Theory and Methods

An optimized pTx-CEST presaturation scheme based on precomputed universal pulses was designed. The optimization was performed by minimizing the L2-norm between the effective B+1 RMS and a given target while imposing energy constraints under virtual observation points (VOPs) supervision. The proposed method was evaluated through simulations and experimentally, both in vitro, on a realistic human head phantom, and in vivo, on healthy volunteers. The results were compared with circular polarization (CP) presaturation and other pTx approaches previously proposed. All experiments were conducted on a 7 T MRI scanner using a commercial 8Tx/32Rx head coil.

Results

The simulations show that the proposed pTx strategy boosted with VOPs is superior to the CP mode and existent pTx approaches. While the best results are obtained with subject specific pulses, the gain provided by the use of VOPs renders the universal pulses superior to tailored pulses optimized under vendor provided Specific Absorption Rate (SAR) management. In the phantom, the glucose MTRasym map was significantly more homogeneous than with CP (root mean square error [RMSE] 17% vs. 30%). The efficiency of the method for in vivo hydroxyl, glutamate and rNOE weighted CEST acquisitions was also demonstrated.

Conclusion

The use of a pTx presaturation scheme based on universal pulses optimized under VOP SAR management is significantly benefiting CEST imaging at high magnetic field.