Evaluating efficient SENSE algorithms to deblur spiral MRI with fat/water separation

link to paper

Evaluating efficient SENSE algorithms to deblur spiral MRI with fat/water separation

Tzu Cheng Chao, Xi Peng, Dinghui Wang, James G. Pipe

Abstract

Purpose

The combination of SENSE and spiral imaging with fat/water separation enables high temporal efficiency. However, the corresponding computation increases due to the blurring/deblurring operation across the multi-channel data. This study presents two alternative models to simplify computational complexity in the original full model (model 1). The performances of the models are evaluated in terms of the computation time and reconstruction error.

Methods

Two approximated spiral MRI reconstruction models were proposed: the comprehensive blurring before coil operation (model 2) and the regional blurring before coil operation (model 3), respectively, by altering the order of coil-sensitivity encoding process to distribute signals among the multi-channel coils. Four subjects were recruited for scanning both fully sampled T1- and T2-weighted brain image data with simulated undersampling for testing the computational efficiency and accuracy on the approximation models.

Results

Based on the examples, the computation time can be reduced to 31%–47% using model 2, and to 39%–56% using model 3. The quality of the water image remains unchanged among the three models, whereas the primary difference in image quality is in the fat channel. The fat images from model 3 are consistent with those from model 1, but those from model 2 have higher normalized error, differing by up to 4.8%.

Conclusion

Model 2 provides the fastest computation but exhibits higher error in the fat channel, particularly in the high field and with long acquisition window. Model 3, an abridged alternative, is also faster than the full model and can maintain high accuracy in reconstruction.