Fat mitigation strategies to improve image quality of radial 4D flow MRI in obese subjects

link to paper

Fat mitigation strategies to improve image quality of radial 4D flow MRI in obese subjects

A. M. K. Muntasir Shamim, Nikolaos Panagiotopoulos, Alma Spahic, David T. Harris, Alejandro Roldán-Alzate, Oliver Wieben, Scott B. Reeder, Thekla Helene Oechtering, Kevin M. Johnson

Abstract

Purpose

This study addresses the challenges in obtaining abdominal 4D flow MRI of obese patients. We aimed to evaluate spectral saturation and inner volume excitation as methods to mitigating artifacts originating from adipose signals, with the goal of enhancing image quality and improving quantification.

Methods

Radial 4D flow MRI acquisitions with fat mitigation (inner volume excitation [IVE] and intermittent fat saturation [FS]) were compared to a standard slab selective excitation (SSE) in a test–retest study of 15 obese participants. IVE selectively excited a cylindrical region of interest, avoiding contamination from peripheral adipose tissue, while FS globally suppressed fat based on spectral selection. Acquisitions were evaluated qualitatively based on expert ratings and quantitatively based on conservation of mass, test–retest repeatability, and a divergence free quality metric. Errors were evaluated statistically using the absolute and relative errors, regression, and Bland–Altman analysis.

Results

IVE demonstrated superior performance quantitatively in the conservation of mass analysis in the portal vein, with higher correlation and lower bias in regression analysis. IVE also produced flow fields with the lowest divergence error and was rated best in overall image quality, delineating small vessels, and producing the least streaking artifacts. Evaluation results did not differ significantly between FS and SSE. Test–retest reproducibility was similarly high for all sequences, with data suggesting biological variations dominate the technical variability.

Conclusion

IVE improved hemodynamic assessment of radial 4D flow MRI in the abdomen of obese participants while FS did not lead to significant improvements in image quality or flow metrics.