Head-and-neck multichannel B1+ mapping and RF shimming of the carotid arteries using a 7T parallel-transmit head coil

link to paper

Head-and-neck multichannel B1+ mapping and RF shimming of the carotid arteries using a 7T parallel-transmit head coil

Matthijs H. S. de Buck, James L. Kent, Peter Jezzard, Aaron T. Hess

Abstract

Purpose

Neurovascular MRI suffers from a rapid drop in B1+ into the neck when using transmit head coils at 7 T. One solution to improving B1+ magnitude in the major feeding arteries in the neck is to use custom RF shims on parallel-transmit head coils. However, calculating such shims requires robust multichannel B1+ maps in both the head and the neck, which is challenging due to low RF penetration into the neck, limited dynamic range of multichannel B1+ mapping techniques, and B0 sensitivity. We therefore sought a robust, large-dynamic-range, parallel-transmit field mapping protocol and tested whether RF shimming can improve carotid artery B1+ magnitude in practice.

Methods

A pipeline is presented that combines B1+ mapping data acquired using circularly polarized (CP) and CP2-mode RF shims at multiple voltages. The pipeline was evaluated by comparing the predicted and measured B1+ for multiple random transmit shims, and by assessing the ability of RF shimming to increase B1+ in the carotid arteries.

Results

The proposed method achieved good agreement between predicted and measured B1+ in both the head and the neck. The B1+ magnitude in the carotid arteries can be increased by 43% using tailored RF shims or by 37% using universal RF shims, while also improving the RF homogeneity compared with CP mode.

Conclusion

B1+ in the neck can be increased using RF shims calculated from multichannel B1+ maps in both the head and the neck. This can be achieved using universal phase-only RF shims, facilitating easy implementation in existing sequences.