Hybrid adiabatic pulse with asymmetry (HAPY): An asymmetric adiabatic pulse with an application in pulsed arterial spin labeling at 7T

link to paper

Hybrid adiabatic pulse with asymmetry (HAPY): An asymmetric adiabatic pulse with an application in pulsed arterial spin labeling at 7T

Didi Chi, Yasmin Blunck, Rebecca Glarin, Catherine E. Davey, Xianyi Zhang, Daniel Stäb, Josef Pfeuffer, Leigh A. Johnston, Jin Jin

Abstract

Purpose

A new class of asymmetric adiabatic radiofrequency (RF) pulses, Hybrid Adiabatic Pulse with asYmmetry (HAPY), is designed to be used as the labeling pulse for Pulsed Arterial Spin labeling (PASL) at 7T to reduce overall specific absorption rate (SAR) while maintaining high labeling efficiency with B0 and B1+ inhomogeneities.

Methods

Realistic Δ⁢B0 and B1+ distributions were extracted from multiple in vivo scans. The proposed class of asymmetric pulses was parameterized and optimized considering these conditions. Simulation and phantoms experiments were performed to compare the optimized pulses with HS-3, GOIA, and trFOCI pulses. In vivo experiments were conducted to demonstrate the application of HAPY in PICORE PASL at 7T, compared with the GOIA and trFOCI pulses.

Results

HAPYs with different amounts of pulse energy reduction are obtained by the proposed optimization framework. Both simulation and phantom experiments demonstrate that HAPY achieves high labeling efficiency and high selectivity along the critical side despite B0 off-resonance and low B1+ amplitude. In vivo experiments reveal that HAPY is able to generate robust perfusion signal with less overall SAR or shorter pulse repetition time, compared to the GOIA and trFOCI pulses.

Conclusion

The HAPY class of pulses, obtained via systematic optimization tailored to the application of PASL at 7T, reduces power deposition without affecting labeling efficiency, which provides a prospect of further exploiting the benefits of ultra-high field in ASL.