Improved accuracy and precision with three-parameter simultaneous myocardial T1 and T2 mapping using multiparametric SASHA

link to paper

Improved accuracy and precision with three-parameter simultaneous myocardial T1 and T2 mapping using multiparametric SASHA

Kelvin Chow, Genevieve Hayes, Jacqueline A. Flewitt, Patricia Feuchter, Carmen Lydell, Andrew Howarth, Joseph J. Pagano, Richard B. Thompson, Peter Kellman, James A. White

Abstract

Purpose

To develop and validate a three-parameter model for improved precision multiparametric SAturation-recovery single-SHot Acquisition (mSASHA) cardiac T1 and T2 mapping with high accuracy in a single breath-hold.

Methods

The mSASHA acquisition consists of nine images of variable saturation recovery and T2 preparation in 11 heartbeats with T1 and T2 values calculated using a three-parameter model. It was validated in simulations and phantoms at 3 T with comparison to a four-parameter joint T1-T2 technique. The mSASHA acquisition was compared with MOLLI, SASHA, and T2-prepared balanced SSFP in 10 volunteers.

Results

The mSASHA technique had high accuracy in phantoms compared to spin echo, with −0.2 ± 0.3% T1 error and −2.4 ± 1.3% T2 error. The mSASHA coefficient of variation in phantoms for T1 was similar to MOLLI (0.7 ± 0.2% for both) and T2-prepared balanced SSFP for T2 (1.3 ± 0.7% vs 1.4 ± 0.3%, adjusted p > .05 for both). In simulations, three-parameter mSASHA had higher precision than four-parameter joint T1-T2 for both T1 and T2 (46% and 11% reductions in T1 and T2 interquartile range for native myocardium). In vivo myocardial mSASHA T1 was similar to SASHA (1523 ± 18 ms vs 1520 ± 18 ms) with similar coefficient of variation to both MOLLI and SASHA (3.3 ± 0.6% vs 3.1 ± 0.6% and 3.3 ± 0.5% respectively, adjusted p > .05 for all). Myocardial mSASHA T2 was 37.1 ± 1.1 ms with similar precision to T2-prepared balanced SSFP (6.7 ± 1.7% vs 6.0 ± 1.6%, adjusted p > .05).

Conclusion

Three-parameter mSASHA provides high-accuracy cardiac T1 and T2 quantification in a single breath-hold with similar precision to MOLLI and T2-prepared balanced SSFP. Further study is required to both establish normative values and demonstrate clinical utility in patient populations.