Referenceless Nyquist ghost correction outperforms standard navigator-based method and improves efficiency of in vivo diffusion tensor cardiovascular magnetic resonance

link to paper

Referenceless Nyquist ghost correction outperforms standard navigator-based method and improves efficiency of in vivo diffusion tensor cardiovascular magnetic resonance

Zimu Huo, Ke Wen, Yaqing Luo, Radhouene Neji, Karl P. Kunze, Pedro F. Ferreira, Dudley J. Pennell, Andrew D. Scott, Sonia Nielles-Vallespin

Abstract

Purpose

The study aims to assess the potential of referenceless methods of EPI ghost correction to accelerate the acquisition of in vivo diffusion tensor cardiovascular magnetic resonance (DT-CMR) data using both computational simulations and data from in vivo experiments.

Methods

Three referenceless EPI ghost correction methods were evaluated on mid-ventricular short axis DT-CMR spin echo and STEAM datasets from 20 healthy subjects at 3T. The reduced field of view excitation technique was used to automatically quantify the Nyquist ghosts, and DT-CMR images were fit to a linear ghost model for correction.

Results

Numerical simulation showed the singular value decomposition (SVD) method is the least sensitive to noise, followed by Ghost/Object method and entropy-based method. In vivo experiments showed significant ghost reduction for all correction methods, with referenceless methods outperforming navigator methods for both spin echo and STEAM sequences at b = 32, 150, 450, and 600 smm−2. It is worth noting that as the strength of the diffusion encoding increases, the performance gap between the referenceless method and the navigator-based method diminishes.

Conclusion

Referenceless ghost correction effectively reduces Nyquist ghost in DT-CMR data, showing promise for enhancing the accuracy and efficiency of measurements in clinical practice without the need for any additional reference scans.