Simultaneous Optimization of MP2RAGE T1-weighted (UNI) and FLuid And White matter Suppression (FLAWS) brain images at 7T using Extended Phase Graph (EPG) Simulations

link to paper

Simultaneous Optimization of MP2RAGE T1-weighted (UNI) and FLuid And White matter Suppression (FLAWS) brain images at 7T using Extended Phase Graph (EPG) Simulations

Ayşe Sıla Dokumacı, Fraser R. Aitken, Jan Sedlacik, Pip Bridgen, Raphael Tomi-Tricot, Ronald Mooiweer, Katy Vecchiato, Tom Wilkinson, Chiara Casella, Sharon Giles, Joseph V. Hajnal, Shaihan J. Malik, Jonathan O’Muircheartaigh, David W. Carmichael

Purpose

The MP2RAGE sequence is typically optimized for either T1-weighted uniform image (UNI) or gray matter–dominant fluid and white matter suppression (FLAWS) contrast images. Here, the purpose was to optimize an MP2RAGE protocol at 7 Tesla to provide UNI and FLAWS images simultaneously in a clinically applicable acquisition time at <0.7 mm isotropic resolution.

Methods

Using the extended phase graph formalism, the signal evolution of the MP2RAGE sequence was simulated incorporating T2 relaxation, diffusion, RF spoiling, and B1+ variability. Flip angles and TI were optimized at different TRs (TRMP2RAGE) to produce an optimal contrast-to-noise ratio for UNI and FLAWS images. Simulation results were validated by comparison to MP2RAGE brain scans of 5 healthy subjects, and a final protocol at TRMP2RAGE = 4000 ms was applied in 19 subjects aged 8–62 years with and without epilepsy.

Results

FLAWS contrast images could be obtained while maintaining >85% of the optimal UNI contrast-to-noise ratio. Using TI1/TI2/TRMP2RAGE of 650/2280/4000 ms, 6/8 partial Fourier in the inner phase-encoding direction, and GRAPPA factor = 4 in the other, images with 0.65 mm isotropic resolution were produced in <7.5 min. The contrast-to-noise ratio was around 20% smaller at TRMP2RAGE = 4000 ms compared to that at TRMP2RAGE = 5000 ms; however, the 20% shorter duration makes TRMP2RAGE = 4000 ms a good candidate for clinical applications example, pediatrics.

Conclusion

FLAWS and UNI images could be obtained in a single scan with 0.65 mm isotropic resolution, providing a set of high-contrast images and full brain coverage in a clinically applicable scan time. Images with excellent anatomical detail were demonstrated over a wide age range using the optimized parameter set.