Ultrafast Z-spectroscopic imaging in vivo at 3T using through-slice spectral encoding (TS-UFZ)
Chongxue Bie, Peter C. M. van Zijl, Deng Mao, Nirbhay N. Yadav
Abstract
Purpose
Acquisition of high-resolution Z-spectra for CEST or magnetization transfer contrast (MTC) MRI requires excessive scan times. Ultrafast Z-spectroscopy (UFZ) has been proposed to address this; however, the quality of in vivo UFZ spectra has been insufficient. Here, we present a simple approach to improve this.
Theory and Methods
UFZ imaging acquires full Z-spectra by encoding the spectral dimension spatially via a gradient applied concurrently with the RF saturation pulse. Different from previous implementations, both this saturation gradient and its readout were applied in the slice direction, resulting in a relatively uniform voxel composition. Phase-encoding was applied in both in-plane directions, allowing additional under-sampling and acceleration.
Results
In phantoms, UFZ imaging with through-slice Z-spectral encoding (TS-UFZ) provided Z-spectra of salicylic acid and egg white in excellent agreement with conventional acquisitions. In vivo brain Z-spectra were influenced by flow through the imaging slice which affected the Z-spectral baseline. Still, CEST signals could be quantified after baseline fitting and mapping the residual CEST signal. Amide proton transfer (APT) contrast intensities obtained by TS-UFZ were on the same order of magnitude as conventional CEST but with different contrast across slice which likely is a result of different tissue regions contributing.
Conclusion
TS-UFZ approach improves signal stability and spectral uniformity over previous implementations and allows high spectral-resolution imaging of saturation transfer effects in the human brain at 3T. This implementation allows for further acceleration by reducing phase encoding steps and thus opens up the possibility of mapping dynamic CEST signals in vivo with a practical temporal resolution.