Wirelessly interfacing sensor-equipped implants and MR scanners for improved safety and imaging

link to paper

Wirelessly interfacing sensor-equipped implants and MR scanners for improved safety and imaging

Berk Silemek, Frank Seifert, Johannes Petzold, Rüdiger Brühl, Bernd Ittermann, Lukas Winter

Abstract

Purpose

To investigate a novel reduced RF heating method for imaging in the presence of active implanted medical devices (AIMDs) which employs a sensor-equipped implant that provides wireless feedback.

Methods

The implant, consisting of a generator case and a lead, measures RF-induced E-fields at the implant tip using a simple sensor in the generator case and transmits these values wirelessly to the MR scanner. Based on the sensor signal alone, parallel transmission (pTx) excitation vectors were calculated to suppress tip heating and maintain image quality. A sensor-based imaging metric was introduced to assess the image quality. The methodology was studied at 7T in testbed experiments, and at a 3T scanner in an ASTM phantom containing AIMDs instrumented with six realistic deep brain stimulation (DBS) lead configurations adapted from patients.

Results

The implant successfully measured RF-induced E-fields (Pearson correlation coefficient squared [R 2] = 0.93) and temperature rises (R 2 = 0.95) at the implant tip. The implant acquired the relevant data needed to calculate the pTx excitation vectors and transmitted them wirelessly to the MR scanner within a single shot RF sequence (<60 ms). Temperature rises for six realistic DBS lead configurations were reduced to 0.03–0.14 K for heating suppression modes compared to 0.52–3.33 K for the worst-case heating, while imaging quality remained comparable (five of six lead imaging scores were ≥0.80/1.00) to conventional circular polarization (CP) images.

Conclusion

Implants with sensors that can communicate with an MR scanner can substantially improve safety for patients in a fast and automated manner, easing the current burden for MR personnel.